Laser Safety Specifications for FARO Laser Line Probes (LLPs)
Laser Safety
All FARO® Laser Line Probe models produce a visible laser beam. The laser is CDRH Class II and IEC Class 2 power level; however operation of the device requires the operator to avoid direct exposure to eyes at all times. You should avoid direct exposure to your eye at all times. Although the human blink reaction to bright light provides a natural mechanism of protection to this visible laser beam, use caution while handling the LLP by pointing it at only the target object.
LLP Technical Specification Table
Laser Line Probe Model | Laser Color / Class (CDRH)* | MPE** | NHZ** | NOHD** | Power Output | Wave Length | Max. Scan Rate |
FAROBlu® xR, xP, xS |
Blue / Class II | 2.55mW/cm2 2.55mW/cm2 |
0.0m 0.0m |
2.41m 1.64m |
<1mW | 450 nm 635 nm (red crosshairs) |
300frames/second 300fps x 4,000points/line = 1,200,000points/sec Windowing enabled: Max up to 600 frames/second 600fps X 2,000points/line = 1,200,000 points/sec |
FAROBlu® xE | Blue / Class II | 2.55mW/cm2 2.55mW/cm2 |
0.0m 0.0m |
2.41m 1.64m |
200frames/second 200fps x 2,000points/line = 400,000points/sec Windowing enabled: 200frames/second 200fps x 2,000points/line = 400,000points/sec |
Prizm™ | Green / Class II | not available | <1mW | 520 nm 635 nm (red crosshairs) |
Color: 120frames/second 120fps x 2,000points/line = 240,000points/sec Monochrome: 300frames/second 300fps x 2,000points/line = 600,000points/sec |
V6 FAROBlu® HD/SD | Blue / Class II | 2.55mW/cm2 2.55mW/cm2 |
0.0m 0.0m |
2.41m 1.64m |
HD: 300frames/second 300fps x 2,000points/line = 600,000points/sec SD: 120frames/second 120fps x 1,000points/line = 120,000points/sec |
V6 (HD) | Blue / Class II M | 2.55mW/cm2 2.55mW/cm2 |
0.0m 0.0m |
2.41m 1.64m |
280 frames/second 280fps x 2,000 points/line = 560,000 points/sec |
V4 (LLP) / V5 (ES) | Red / Class II M | not available | <1mW | 660 nm | 60 frames/second 60fps x 752points/line = 45,120 points/sec |
V3 | Red / Class II M | <1mW | 660 nm | 30 frames/second 30fps x 640 points/line = 19,200 points/sec |
V2 / V1 | Red / Class II M | <1mW | 660 nm | 30 frames/second 30fps x 640 points/line = 19,200 points/sec |
*The Center for Devices and Radiological Health (CDRH).
**For more information on MPE, NHZ, and NOHD values, see: Laser Safety Calculation Guide
***xE (not available in US/Canada)
llp, safety, class, eyes